Marchrider
Well-Known Member
They do exist, and they do need a special circuit (although other things can also use that circuit, it isn't the case that the socket can only be used for x-ray machines, but that x-ray machines can only use the marked sockets).
Https://electrical.theiet.org/media/2233/gn7-5th-edition-ch09_reprintv2.pdf (Page 82)
system. (b) Designers should take note of the inherent high protective conductor (earth leakage) currents associated with a variety of electronic equipment. Examples of these are information technology equipment and mobile X-ray equipment, which can produce a higher allowable earth leakage current than normal medical electrical equipment.
And
For example, a typical X-ray generator uses a rectifier circuit to generate a d.c. intermediate voltage. Due to this, equipment manufacturers often advise the use of a Type B RCD.
thanks for that, I sort of understand that - plugged into a standard circuit it could easily trip the RCD
I did ask one of the nurses but she said that all she knew was it mustn't be used for anything else; this explanation seemed crazy, if it was a different voltage or something, why use a standard 3 pin socket that anything else could be plugged into - but considering the madness in how hospitals are run then, who knows.